Describe the autoclave with its principle, types, uses and sterilisation control.
Principle
The autoclave sterilises by saturated steam under pressure, which kills by coagulation and denaturation of microbial protein.
Water boils at 100°C at atmospheric pressure → Increasing the pressure raises the boiling point → At 15 lb/in² (1 atmosphere above ambient), steam reaches 121°C → Steam condenses on the cooler load, releasing a large amount of latent heat and wetting the surface → Moisture and heat together denature and coagulate protein → Holding for 15 minutes at 121°C kills all organisms including spores
| Temperature | Pressure | Holding time |
|---|---|---|
| 121°C | 15 lb/in² | 15 minutes — the standard cycle |
| 126°C | 20 lb/in² | 10 minutes |
| 134°C | 30 lb/in² | 3 minutes — the "flash" cycle; and 18 minutes for prion inactivation |
CLINICAL PEARL
Complete removal of air is the single most important requirement, and the commonest cause of failure. Air is a poor conductor of heat and does not mix freely with steam, so a trapped air pocket sits at a lower temperature than the chamber gauge reports. Instruments inside it are not sterilised, while every physical indicator says the cycle was correct. This is precisely why the BOWIE-DICK test, which checks air removal and steam penetration, is performed daily on vacuum autoclaves.
Types of Autoclave
- Simple (pressure cooker) type — downward displacement, in which steam entering at the top pushes the denser air out through a discharge tap at the bottom; suitable for unwrapped solid instruments and fluids
- Gravity displacement — the standard hospital jacketed autoclave, working on the same principle
- Pre-vacuum (high-vacuum) type — a pump evacuates the chamber before steam is admitted, so air removal is virtually complete and penetration into porous loads is immediate. It allows shorter cycles at 134°C and is essential for wrapped packs, dressings, linen and porous loads
Uses and Limitations
- Uses — surgical instruments, dressings, gowns and linen, gloves, rubber and suitable plastics, culture media, aqueous solutions, and the decontamination of infectious laboratory waste before disposal
- Not suitable for — oils, greases and powders (steam cannot penetrate them, so a hot air oven is used); heat-labile plastics, endoscopes and prosthetic materials (ethylene oxide or hydrogen peroxide plasma); and sharp cutting edges, which are blunted
- Practical requirements — do not overload; leave space between packs; open the taps or lids of containers so steam can enter; and begin timing only when the whole load has reached temperature, not when the gauge does
- Fluids must be cooled slowly to avoid flash boiling and bottle explosion when the door is opened
Sterilisation Control
- Physical — thermometer, pressure gauge, timer and a printed cycle chart; these confirm the machine's conditions but not the centre of the load
- Chemical — autoclave tape (shows exposure only, not sterility), BROWNE tubes (red to green), and the BOWIE-DICK test for air removal, done as the first cycle of each day
- Biological — the definitive control: spore strips of Geobacillus stearothermophilus, which are the most heat-resistant organisms known and germinate at 55 to 60°C if the cycle failed. If these spores are dead, everything else is
- Records must be kept for every cycle, with the load identified, so that a failed cycle can be traced and the instruments recalled
Applied Aspects
- Autoclave tape proves exposure, not sterility — relying on it alone is the commonest false reassurance in a small hospital or clinic
- Run a biological indicator regularly (weekly, and with every implant load), and act on a failure by recalling and reprocessing
- Clean instruments before autoclaving; blood and tissue shield organisms from steam and may bake on permanently
- Autoclave infectious laboratory and hospital waste before disposal, in line with the Biomedical Waste Management Rules
- Do not use a "flash" cycle as routine practice for convenience; it is intended for an urgently needed unwrapped instrument, and packs sterilised this way cannot be stored
- Where autoclaves are unreliable or power is intermittent, as in some peripheral Indian facilities, this must be recognised and addressed rather than assumed away — unsterile instruments transmit hepatitis B, C and HIV
All questions in General Microbiology & Sterilisation
- Q1Classify micro-organisms. Describe the bacterial cell wall and other structures with their applied importance. ★
- Q2Define sterilisation and disinfection. Describe the physical and chemical methods with their control. ★
- Q3Describe bacterial nutrition and growth. Classify culture media with examples and describe anaerobic culture. ★
- Q4Describe bacterial genetics including the mechanisms of gene transfer and the basis of drug resistance. ★
- Q5Describe the normal flora and the host-parasite relationship including virulence factors and host defences. ★
- SN1Describe the autoclave with its principle, types, uses and sterilisation control. ★
- SN2Describe the bacterial spore with its structure, basis of resistance and medical importance. ★
- SN3Describe the Gram stain and the Ziehl-Neelsen stain with their principles and applied importance. ★
- SN4Describe the methods of anaerobic culture and the collection of specimens for anaerobic infection. ★
- SN5Describe the biosafety levels and the management of biomedical waste. ★
- SN6Describe the bacteriophage with its life cycles and importance in microbiology. ★
- SN7Describe the collection of specimens and the methods of antimicrobial susceptibility testing.