Describe quality control in the clinical biochemistry laboratory.
Definition
Quality control = the system of procedures that ensures laboratory results are reliable, reproducible and clinically usable.
| Term | Meaning |
|---|---|
| Accuracy | Closeness to the true value; affected by systematic error (bias) |
| Precision | Reproducibility on repeated measurement; affected by random error |
| Sensitivity | Ability to detect true positives; a sensitive test is good for ruling out |
| Specificity | Ability to identify true negatives; a specific test is good for ruling IN |
| Coefficient of variation | (SD ÷ mean) × 100 — the usual measure of precision |
- Mnemonic — SnNOut (a Sensitive test, if Negative, rules Out) and SpPIn (a Specific test, if Positive, rules In)
- A method can be precise but inaccurate — consistently wrong. Both are needed
Types of Quality Control
| Type | Description | Purpose |
|---|---|---|
| Internal (IQC) | Control sera of known value run with each batch, daily | Monitors precision day to day |
| External (EQAS) | Unknown samples sent by a central agency and compared between laboratories | Monitors accuracy against peers |
The Levey–jennings Chart
- Control values are plotted daily against the mean ± 1, 2 and 3 standard deviations
- 95% of results should fall within ± 2 SD
- Patterns are interpreted by the Westgard rules
| Westgard rule | Meaning | Suggests |
|---|---|---|
| 12s | One value beyond 2 SD | Warning only — expected in 1 of 20 runs |
| 13s | One value beyond 3 SD | Reject — random error |
| 22s | Two consecutive beyond the same 2 SD limit | Reject — systematic error |
| R4s | Two values differing by more than 4 SD | Random error |
| 10x | Ten consecutive on the same side of the mean | Shift — new reagent lot, recalibration |
| 7T | Seven consecutive rising or falling | Trend — deteriorating reagent, ageing lamp, electrode drift |
CLINICAL PEARL
A shift is abrupt; a trend is gradual. The distinction points to different causes — a shift usually follows a discrete event such as a new reagent lot or recalibration, while a trend suggests something deteriorating steadily.
Sources of Error
| Phase | Share of errors | Examples |
|---|---|---|
| Pre-analytical | 60–70% — the largest | Wrong patient or label, haemolysis, wrong tube, tourniquet too long, delay in transport, non-fasting sample |
| Analytical | 10–15% | Instrument fault, reagent deterioration, calibration error, interference |
| Post-analytical | 15–20% | Transcription error, delayed reporting, failure to communicate a critical value |
- Most laboratory error occurs before the sample reaches the analyser — which is why sample collection deserves as much attention as the assay
- Haemolysis falsely raises potassium, LDH, ast and phosphate — the commonest single interference
Applied Aspects
- Critical (panic) values must be telephoned immediately — potassium above 6.5, glucose below 40, calcium above 13, and a positive blood culture
- Always question a result that does not fit the patient; repeat before acting on a surprising value
- Reference ranges are population-specific, defined as the central 95%, which means 1 in 20 healthy people falls outside by definition — the reason a panel of 20 tests so often shows one "abnormal" result
All questions in Clinical & Applied Biochemistry
- Q1Classify the liver function tests. Describe the biochemical findings in the three types of jaundice. ★
- Q2Classify the renal function tests. Describe the assessment of glomerular filtration rate and the tests of tubular function. ★
- Q3Describe the biosynthesis of haem and its regulation. Classify the porphyrias and add a note on lead poisoning. ★
- Q4Classify the acid–base disorders. Describe their causes, compensation and the use of the anion gap. ★
- Q5Describe the formation, transport, conjugation and excretion of bilirubin. Classify jaundice and add a note on neonatal jaundice. ★
- SN1Classify the tumour markers with their clinical applications and limitations. ★
- SN2Describe the normal composition of cerebrospinal fluid and the changes in meningitis. ★
- SN3Describe the thyroid function tests and their interpretation. ★
- SN4Describe quality control in the clinical biochemistry laboratory. ★
- SN5Describe automation in the clinical biochemistry laboratory with its advantages and limitations.
- SN6Describe the principles and applications of therapeutic drug monitoring.
- SN7Describe the principles of biochemical screening with examples in the newborn and in pregnancy. ★